IW9EZO said "Probable not precise, I'm new in the measures."
OK, taking the risk I'm telling nothing new, perhaps a 'short cut' advise (assuming the directional coupler is ok):
- Leave sweep output on 0 dBm (1 mW)
- Disconnect the directional coupler from the antenna
- Place a proper 50Ω load as 'antenna'
- Measure the response at 2400 MHz
(if it's a good 50Ω load you should see a (very) low response)
- Disconnect the load
- Measure the response at 2400 MHz
(should be maximum because all power is reflected)
(normally one should also make a short, but I leave that out ; -)
Now you have the difference between an 'ideal' 50Ω load and no load.
- Connect the antenna and measure the level at 2400 MHz.
The RL Is the difference between the level of the 'open situation' and the level you measure with the helix.
To see if there is a 'dip' try to increase the span to e.g. 1000 MHz